EDA test - needs vigorous activity

Connection, ECG, EMG, EDA, etc.
mvallance
Posts: 5
Joined: Tue Oct 24, 2017 4:44 am

EDA test - needs vigorous activity

Postby mvallance » Mon Oct 30, 2017 7:22 am

When doing a EDA test in our constructed Bitalino device (Revolution BLE/ Freestyle), we obtain data in OpenSignals. However, I have to do a vigorous activity to make the graph appear as peaks and valleys. When I stop, the graph reverts to a relatively flat line.
The point is, the data acquisition appears to be quite extreme: if I do a relatively medium activity (like walking on the spot) I get no data shown.

We have connected/reconnected device, and even re-soldered our board. We use Bitalino pre-gelled electrodes.

Can you advise where I should be troubleshooting? Thanks.

BITalino
Site Admin
Posts: 567
Joined: Tue Aug 27, 2013 3:47 pm

Re: EDA test - needs vigorous activity

Postby BITalino » Mon Oct 30, 2017 7:56 am

Hi there,

First of all we'd like to thank you for your preference for BITalino. How and where are you placing the electrodes?

We have a few guidelines on the following thread of our forum and also on the data sheet:
http://forum.bitalino.com/viewtopic.php?t=135
http://bitalino.com/datasheets/REVOLUTION_EDA_Sensor_Datasheet.pdf

Best regards,
The BITalino Team

mvallance
Posts: 5
Joined: Tue Oct 24, 2017 4:44 am

Re: EDA test - needs vigorous activity

Postby mvallance » Mon Oct 30, 2017 10:52 am

That's a good starting point. Thank you. We had success with ECG sensors. See image attached.

ECG_signal1.jpg
ECG_signal1.jpg (161.04 KiB) Viewed 497 times


Though now we need to try EDA sensors.

BITalino
Site Admin
Posts: 567
Joined: Tue Aug 27, 2013 3:47 pm

Re: EDA test - needs vigorous activity

Postby BITalino » Mon Oct 30, 2017 12:01 pm

If you have the chance, please send us a couple of photos of your existing setup... the ECG signal should look a bit differently also.

Best regards,
The BITalino Team

mvallance
Posts: 5
Joined: Tue Oct 24, 2017 4:44 am

Re: EDA test - needs vigorous activity

Postby mvallance » Tue Oct 31, 2017 12:24 am

This is the OpenSignals connection.

opensignal_conn1.PNG
opensignal_conn1.PNG (112.46 KiB) Viewed 491 times


Here are 2 images of our current circuit.

circuit1a.jpg
circuit1a.jpg (106.71 KiB) Viewed 491 times

-

circuit1b.jpg
circuit1b.jpg (126.78 KiB) Viewed 491 times



Thank you.

BITalino
Site Admin
Posts: 567
Joined: Tue Aug 27, 2013 3:47 pm

Re: EDA test - needs vigorous activity

Postby BITalino » Tue Oct 31, 2017 12:36 am

Looks great... thanks for sharing.

Best regards,
The BITalino Team

mvallance
Posts: 5
Joined: Tue Oct 24, 2017 4:44 am

Re: EDA test - needs vigorous activity

Postby mvallance » Fri Nov 03, 2017 12:01 am

the ECG signal should look a bit differently also.

In what way should it be different?

Thanks.

BITalino
Site Admin
Posts: 567
Joined: Tue Aug 27, 2013 3:47 pm

Re: EDA test - needs vigorous activity

Postby BITalino » Fri Nov 03, 2017 6:57 am

The amplitude of the peaks should be higher... you can see an example here:
http://bitalino.com/datasheets/REVOLUTION_ECG_Sensor_Datasheet.pdf

Nevertheless, it may also be related with the scale of the signal.

Best regards,
The BITalino Team

mvallance
Posts: 5
Joined: Tue Oct 24, 2017 4:44 am

Re: EDA test - needs vigorous activity

Postby mvallance » Sat Nov 04, 2017 1:26 am

We'll check this. Thank you.